View Full Version : how to calibrate ppm on hanna hi98129



Felixthecat
10-15-2009, 10:53 PM
do i use the 1413us/cm solution or the 1500 ppm solution to calibrated

altaholic
10-16-2009, 12:38 AM
With those Hanna meters make sure you use the right calibration solution, only the packets that come with it. GH has a ppm calibration fluid and I guess it doesnt work with them. I use the 1500 ppm solution and its fine. It does take a while so leave it in there for a couple of minutes.

Felixthecat
10-22-2009, 05:51 PM
is it ok with the genesis calibration solution cause i order the hanna one but dude from the store gave me a dual package genesis, one is 1413us/989ppm and the other is 1500ppm/2142us both at .70 conversion factor

DreadedHermie
10-22-2009, 07:22 PM
For TDS measurements, users
can choose a conversion factor
between 0.45 and 1.00 in
increments of 0.01.

Set your meter to "conv 0.70," and those solutions should be fine. They give you two conductivities so you can compare and "balance the results."

Felixthecat
10-23-2009, 06:34 PM
thanx bro i was gonna return that shit and reorder but no need for that so after calibrating my pen it would be excellent everything would be on dot
with perfect PH and right PPM i just cant wait

Felixthecat
10-23-2009, 11:42 PM
hey bro i dont really get what the instruction is telling me, how do you calibrate ppm cause there are two

DreadedHermie
10-24-2009, 12:14 AM
According to the specs, this meter will calibrate itself using a single solution (1413 μS/cm). You can always use the other cal solution as a "check," though.

Is this what your instructions say?

What part don't you understand?

Just put it in "cal" mode and dip it in! :thumbsup: It's automatic!



Calibration procedure

â?¢ From measurement mode, press and hold the /
MODE button until CAL is displayed on the lower
LCD.

â?¢ Release the button and immerse the probe in the
proper calibration solution: HI7031 (1413 μS/cm)
for HI98129

â?¢ Once the calibration has been automatically performed,
the LCD will display OK for 1 second and
the meter will return to normal measurement mode.

Felixthecat
10-24-2009, 02:12 AM
the whole part u just said nah j/k but thanx alot for the help,thanx all you guys:rasta:

Felixthecat
10-24-2009, 02:24 AM
what is the BETA do i change that too are just leave it